identifying secondary peaks
Posted: Wed Feb 22, 2012 9:11 am
hi,
i just want to ask how can i clearly identify a secondary peak in my gradient run?
i am currently doing the assay and impurity test for carbimazole tablets following the procedure as written in the british pharmacopoeia 2009.
my chromatographic comdition are as follows:
i am using waters hplc 2695
column: i have tried to use phenomenex Luna 250 x 4.6 5um, synergi FP 250 x 4.6 4 um. the BP 2009 required a column packed with base deactivated octadecylsilyl sical gel 150 mm
flow rate : 1.5 ml/min i have increased the flow rate because i also increased the length of the column..
the MPA : 5% ACN
MPB :20% ACN
time MPA MPB
0 100 0
10.0 100 0
10.1 0 100
36 0 100
36.1 100 0
46 100 0
i equilibrated the column for 1 hour on MP A
and 15 mins on MPB
i am required to inject
3 standard solutions
1. 0.0025% w/v carbimazole in 5% ACN
2. 0.0005% w/v thiamazole in 5% ACN
3. 0.01% w/v carbimazole and 0.0005% w/v thiamazole in 5% ACN
now here is my problem :
i inject the MPA first the 5% ACN for this is the one i used as a solvent of my samples...
the chromatograph looks clear to me except some baseline change at 14 mins...
when i injected the standard 1 the carbimazole standard i has no peak at the RT of thiamazole but it has a visible peak at 16 mins. the carbimazole peak eluted at around 19 mins.. is this a secondary peak? i am using my pure carbimazole standard here..
then i try to inject standard 2 the thiamazole standard which the standard for the impurity test.
it also have a clear peak at 16 mins, i really dont know where this peak is coming from. the thiamazole eluted at 5 mins..
then for my standard 3 which is for my system suitability, there are 3 peaks, the carbimazole peak ,the thiamazole peak which i think the area is 20x more the area of the standard 2 (Note : the concentration of my standard 2 is the same as this conc.) and my unknown peak at 16 mins..
can the peak at 16 mins be considered as my secondary peaks? it is realy big. both of my standard will fail the impurity test if i consider this as the secondary peak... i dont know how to justify this one..
and also my sample also have this peak at 16 mins..
please help
i have testing my sample for 3 weeks now..
i dont know if i will consider this unkown peak as a secondary peak.... or there is something wrong in the procedure that i do...
thank you so much ....
the MPA : 5% ACN
MPB :20% ACN
time MPA MPB
0 100 0
10.0 100 0
10.1 0 100
36 0 100
36.1 100 0
46 100 0
and 15 mins on MPB
please help