HPLCAddict beat me to the reply while I was composing, but here it is anyway:
I don't understand, the compound is soluble up to 1g/mL, how come it can be trap into the flow cell?
Many compounds can adsorb to stainless steel. The nitric acid "passivation" presumably leaves behind a thin film of oxide (manganese?) which masks the underlying iron.
All of that said, as I re-read your original post, I doubt that the problem is in the flow cell. In fact, if you are seeing a distinct carryover peak for your compound, it is almost certainly *not* the flow cell. You have already exonerated the column, so the remaining possibilities are in the sampler/injector or the transfer lines and fittings between the injector and the column. It *is* possible to diagnose the source (I'll put in a plug for our
HPLC Basics, Equipment, and Troubleshooting course on the web next month!). Here's what you do:
Inject a high-level standard followed by 2-3 blanks.
If the peak areas in the blanks decrease exponentially (e.g., the first blank shows 1% of the standard, the second shows 0.01% -- 1% of 1%, etc.) then the problem is probably "physical" carryover of sample trapped in poorly-flushed void spaces. Check for excess tubing, poorly assembled fittings, incomplete washing of the syringe, etc.
If the peak areas in the blanks decrease more slowly than exponentially (e.g., 0.1%, 0.08%, 0.06%, etc.) then you are probably looking at "chemical" carryover; adsorption of analyte to a surface somewhere in the system. This can be a real headache to track down. Remedies may include flushing with organic solvent, passivating the entire system with dilute nitric acid, or changing tubing from stainless steel to PEEK (or vice versa).
If the peak areas in the blanks stay roughly constant *and* you are running a gradient, the most likely cause is contamination of the "A" reservoir. You can confirm this with the "three blank gradients" test described in the mini-tutorial on gradient baseline issues that I posted at the top of the board (
viewtopic.php?f=1&t=19085)