DB-1 column for test Ethylene Chlorohydrin

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi all,

please advise for test Ethylene chlorohydrin by DB-1 column with many thank

thx a lot
Aimily :lol:
http://www.restek.com/chromatogram/view/GC_EV1169

see peak location for 2-chloroethanol

DB-1 column should be similar in retention times given the diameter and phase thickness.

Rod
Dear Rod,

Many thank for advise :D
these chromatogram is GC or GC-MS?
if i use GC-FID may be not same this chromatogram,right?

thx a lot
Aimily :lol:
"these chromatogram is GC or GC-MS?"

GC

It is not the detector which determine RT, but carrier gas, its flow rate, column temperature, phase film type, its thickness, are the variables involved. There may be and there are variances between columns and even intracolumn which causes RT variations. You are expecting too much to predict with any accuracy the retention time of an analyte on a column.

best wishes

Rod
4 posts Page 1 of 1

Who is online

In total there are 42 users online :: 0 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry