Ghost Peak after Column Washing with 0.1% TFA

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Dear friends,

I was consistently getting 'ghost peak' and suspected contamination in my column {BEH Shield RP-18 (100x2.1) mm, 1.7 micron}.

I washed the column with A 90% to B 90% for a gradient time of ten minutes, in which A is 0.1% TFA in water and B is 0.1% TFA in acetonitrile. Unfortunately, instead of solving the problem, I got three more ghost peaks. They are now appearing even after washing with 10% aqueous ACN followed by 100% ACN.

Please Help!
Run the "three blank gradients" test to see if the cause is contaminated mobile phase. It's described in the "mini tutorial" in the tophread on the board.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry