Agilent 6890
Rtx-5 105m x .25mm x .25um
Analytes: Hydrocarbons from C5-C14(ish)
I've noticed that the major peaks in my chromatograms are shifting up to as much as 0.2 minutes (on a 105m column with about a 1.5hr run time). This, of course, is a problem, but I can't figure out what is causing the shifting. I've considered all the basic causes and ruled them out: tThe column is new (roughly 2 months old), carrier solvent is and has always been DCM, flow rate and oven temp are constant between runs.
I considered that the sample may be too concentrated--each sample is made by adding 5 drops sample, then filling to 1.5mL with DCM...so with each drop being roughly 25mg and the major component at 90% or above, that puts the concentration of that particular component at 100mg/mL (roughly) which is way too concentrated, I know.
My issue with this is that I did 3 runs each using 1 drop, 2 drops, 3 drops, 4 drops, and 5 drops of sample to get a handle on the way the concentration is affecting the shifting of retention times--there was no significant difference between the different samples. I ran these samples on a shorter column though to save time, so perhaps it wasn't an accurate barometer for my problem. But to give you an idea of the repeatability I'm looking for, between each of those runs, the most any peak shifted was 0.004 minutes.
I'm running out of ideas...help!!