Decreasing response of GCMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
My dear friends hello to all of you.
I absorbed one thing during the autotune of my GCMS,the response of the ion is decreasing gradually. Can this will effect my chromatograms,if so than what should i do ?


Chandra
Can you give us more details?

-which ion is decreasing (or are all ions in tune)
-what MS are you using?
-If Agilent, which filament are you using?
-How long are you tuning for? Do you tune multiple times over about an hour?
-Do you have enough PFTBA (FC43) tune mixture?

any other details you can provide?

Ty
~Ty~
This should not directly effect your retention times, just peak TIC's. This is why daily to weekly (pending on how often you run) calibrations and tunings are so important. If you are looking at peaks close to the s/n ration this could become an issue to where the source needs to be cleaned.
3 posts Page 1 of 1

Who is online

In total there are 45 users online :: 0 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 45 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry