I'm running a silica column with pure hexane and not getting any retention for dimethylfuran
or my xylene mixture. Everything comes unretained. I'm using UV detection. I can inject them
separately and see an unretained peak. My mobile phase is 100% hexane but I ramp up to
50% EtOH over 8 minutes. I can inject a blank hexane and see nothing.

Should I expect retention for these components? I can not run reversed phase or GC.
I would expect retention, but how much would depend on the specific surface area of the silica. Another factor is the fact that silica is very touchy about the trace water content of the mobile phase (water adsorbs to and deactivates the silica). The same is true about alcohols, so one possibility is that you are not allowing sufficient re-equilibration time after running the gradient to EtOH (it can take a looooong time).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 578 users online :: 1 registered, 0 hidden and 577 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 577 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry