I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
-
- Posts: 264
- Joined: Sun Nov 14, 2010 2:20 pm
Need more info... but a guess.... impurities are in the solvent you are using?I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
- Karen
-
- Posts: 21
- Joined: Thu Feb 02, 2012 8:40 am
The HS solvent was purified water which was a "Blank" shot too. There was no any impurity peaks in this blank. Any other info would you need pls? Thanks!Need more info... but a guess.... impurities are in the solvent you are using?I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
- Karen
-
- Posts: 53
- Joined: Tue May 26, 2009 8:17 pm
If the system sits idle overnight with no runs, is your contamination higher on the first run of the day?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Was the blank prepared at the same time as the standards? or at a different time.
The blank removes the possibility that the system produced the peaks.
My guess is the atmosphere or the vials were contaminated with the peaks you have found.
As I have seen 'pizza' peaks and 'aftershave' peaks in analyses I have performed, I am always amazed how sensitive HS analysis can be.
best wishes,
Rod
The blank removes the possibility that the system produced the peaks.
My guess is the atmosphere or the vials were contaminated with the peaks you have found.
As I have seen 'pizza' peaks and 'aftershave' peaks in analyses I have performed, I am always amazed how sensitive HS analysis can be.
best wishes,
Rod
-
- Posts: 21
- Joined: Thu Feb 02, 2012 8:40 am
No, it was a consequtive run, there was no stop until the completion.If the system sits idle overnight with no runs, is your contamination higher on the first run of the day?
-
- Posts: 21
- Joined: Thu Feb 02, 2012 8:40 am
Very good questions and guesses. We are going to look into them.Was the blank prepared at the same time as the standards? or at a different time.
The blank removes the possibility that the system produced the peaks.
My guess is the atmosphere or the vials were contaminated with the peaks you have found.
As I have seen 'pizza' peaks and 'aftershave' peaks in analyses I have performed, I am always amazed how sensitive HS analysis can be.
best wishes,
Rod
BTW, what are the "pizza" and "aftershave" peaks please?
Tanks a lot!
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
I did not run mass spec to identify the peaks conclusively.
volatile Oils in the pizza and fragrance components of the aftershave, small carbon number esters and aldehydes, were the most likely candidates.
Rod
ps clean baselines = clean chemist
volatile Oils in the pizza and fragrance components of the aftershave, small carbon number esters and aldehydes, were the most likely candidates.
Rod
ps clean baselines = clean chemist
8 posts
Page 1 of 1
Who is online
In total there are 40 users online :: 0 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 40 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 40 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science