Retention time and system suitability

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Could you please help on defining system suitability criteria regarding retention time? I am used to work upon RSD < 1% regarding 6 replicate injections, but there is no evaluation provided in the current phamacopoeias. I assume it is a rational deviation, but no theoritical information is provided.
The rationale for 1% is essentially "because we can".

If you want to set a looser spec and you have the robustness data to back it up that should be OK (you may have to do some convincing of a reviewer). If you have a complex sample or close potential interfering peaks, then you might have to set your spec a bit tighter. You can't go too much tighter because many HPLC pumps only spec flow accuracy to something like 0.25%
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot], Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry