Analyte loss at injector port?
Posted: Tue Feb 14, 2012 11:23 pm
First and foremost, our setup and relevant info:
Agilent 6850
Rtx-5 105m x .25mm x .25um (bonded polymethylsiloxyl, capillary GC column)
Injector port temp: 160C
Split Ratio: 20:1
Injection Size: 1uL
Constant pressure (44.14psi) - Flow = 2mL/min and Avg. Velocity at 25.
Packed, quartz wool, tapered liner
Analytes: C5 (hydrocarbons)
After pulling vacuum on our sample, we recovered approximately 6% of what we're calling "solvent lights" (meaning low molecular weight analytes that are covered on our chromatographs by the solvent peak--of course, we use a different solvent when looking at these peaks). When we GC the material prior to pulling vacuum, we're seeing it at about 3%. The discrepancy between these numbers has us thinking that we're possibly losing material at the injector port because it is so volatile (b.p.'s range between approximately Isoprene 34C - Cyclopentene 44C).
Thoughts?
Agilent 6850
Rtx-5 105m x .25mm x .25um (bonded polymethylsiloxyl, capillary GC column)
Injector port temp: 160C
Split Ratio: 20:1
Injection Size: 1uL
Constant pressure (44.14psi) - Flow = 2mL/min and Avg. Velocity at 25.
Packed, quartz wool, tapered liner
Analytes: C5 (hydrocarbons)
After pulling vacuum on our sample, we recovered approximately 6% of what we're calling "solvent lights" (meaning low molecular weight analytes that are covered on our chromatographs by the solvent peak--of course, we use a different solvent when looking at these peaks). When we GC the material prior to pulling vacuum, we're seeing it at about 3%. The discrepancy between these numbers has us thinking that we're possibly losing material at the injector port because it is so volatile (b.p.'s range between approximately Isoprene 34C - Cyclopentene 44C).
Thoughts?