A HS-GC Question

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
Need more info... but a guess.... impurities are in the solvent you are using?

- Karen
I ran a series (9) of reference stanards acrossing 100 times concentration range (40-4000ug/mL) on a HS-GC/FID. It was found that the 5 impurity peak areas remianed almost unchanged while the target peak aears increased reasonably proportional (R2=0.9995) with the concentration increasing in each level. Please help to explain why!
Need more info... but a guess.... impurities are in the solvent you are using?

- Karen
The HS solvent was purified water which was a "Blank" shot too. There was no any impurity peaks in this blank. Any other info would you need pls? Thanks!
If the system sits idle overnight with no runs, is your contamination higher on the first run of the day?
Was the blank prepared at the same time as the standards? or at a different time.

The blank removes the possibility that the system produced the peaks.

My guess is the atmosphere or the vials were contaminated with the peaks you have found.

As I have seen 'pizza' peaks and 'aftershave' peaks in analyses I have performed, I am always amazed how sensitive HS analysis can be.

best wishes,

Rod
If the system sits idle overnight with no runs, is your contamination higher on the first run of the day?
No, it was a consequtive run, there was no stop until the completion.
Was the blank prepared at the same time as the standards? or at a different time.

The blank removes the possibility that the system produced the peaks.

My guess is the atmosphere or the vials were contaminated with the peaks you have found.

As I have seen 'pizza' peaks and 'aftershave' peaks in analyses I have performed, I am always amazed how sensitive HS analysis can be.

best wishes,

Rod
Very good questions and guesses. We are going to look into them.

BTW, what are the "pizza" and "aftershave" peaks please?

Tanks a lot!
I did not run mass spec to identify the peaks conclusively.

volatile Oils in the pizza and fragrance components of the aftershave, small carbon number esters and aldehydes, were the most likely candidates.

Rod

ps clean baselines = clean chemist
8 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry