is ELSD similiar to FID in terms of quantification
Posted: Tue Feb 14, 2012 8:48 pm
can I use % analyte =100*(analyte peak/SUM(all peaks)) for ELSD if I only need semi-quantification result.
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
As AA and CPG have suggested, the answer is "maybe".
An ELSD essentially detects "dust" left behind when the solvent is evaporated from the droplets that come out of the nebulizer. To a first approximation, "dust is dust" and the response to each dust particle does not depend on the chemical structure of the non-volatile material. So, there is much less variation in response factors with ELSD than with UV.
That said, the presence of analyte may affect the number and/or size distribution of dust particles formed (e.g., by changing the surface tension of the column effluent or by affecting the rate of evaporation of solvent), so it's unrealistic to expect response factors to be exactly the same even for an isocratic separation (with a gradient of course, droplet formation may change quite a bit as a function of the mobile phase composition).
The relatively narrow (actually, nonexistent!) linear range is, to my mind, much less of an issue; I have no problems with appropriate mathematical transforms -- we do it all the time with UV detectors, which actually measure transmittance and then compute absorbance from that.
Bottom line: for "semi-quantitative" estimate, ELSD area% is probably a reasonable choice (certainly better than UV), but don't expect too much!
Unfortunately, none that I am aware of.Do we have some sort of equation similar to lamb-beer law for ELSD?
Unfortunately, none that I am aware of.Do we have some sort of equation similar to lamb-beer law for ELSD?