is ELSD similiar to FID in terms of quantification

Posted: Tue Feb 14, 2012 8:48 pm
by jiangds06
can I use % analyte =100*(analyte peak/SUM(all peaks)) for ELSD if I only need semi-quantification result.

Re: is ELSD similiar to FID in terms of quantification

Posted: Wed Feb 15, 2012 4:15 am
by AA
ELSD's are non-linear.
In ELSD each compound will generate a different resopnse factor (i.e. for the same mass on column, you can get very different peak heights, areas).

I would hesitate to do what you are planning to do.

Re: is ELSD similiar to FID in terms of quantification

Posted: Wed Feb 15, 2012 3:32 pm
by Consumer Products Guy
ELSD is less linear than UV, linear over a smaller range. But molecules that are different can have closer detector responses.

For example, if a sample contains trace naphthalene, the UV peak could be huge, but just a small amount present, due to its molecular absorptivity. So just realize the limitations of each.

Re: is ELSD similiar to FID in terms of quantification

Posted: Wed Feb 15, 2012 8:41 pm
by tom jupille
As AA and CPG have suggested, the answer is "maybe".

An ELSD essentially detects "dust" left behind when the solvent is evaporated from the droplets that come out of the nebulizer. To a first approximation, "dust is dust" and the response to each dust particle does not depend on the chemical structure of the non-volatile material. So, there is much less variation in response factors with ELSD than with UV.

That said, the presence of analyte may affect the number and/or size distribution of dust particles formed (e.g., by changing the surface tension of the column effluent or by affecting the rate of evaporation of solvent), so it's unrealistic to expect response factors to be exactly the same even for an isocratic separation (with a gradient of course, droplet formation may change quite a bit as a function of the mobile phase composition).

The relatively narrow (actually, nonexistent!) linear range is, to my mind, much less of an issue; I have no problems with appropriate mathematical transforms -- we do it all the time with UV detectors, which actually measure transmittance and then compute absorbance from that.

Bottom line: for "semi-quantitative" estimate, ELSD area% is probably a reasonable choice (certainly better than UV), but don't expect too much! :wink:

Re: is ELSD similiar to FID in terms of quantification

Posted: Tue Feb 21, 2012 3:36 pm
by jiangds06
Wow!
thanks so much. For UV, A =log(I0/I). This is definitely a mathematical transformation.
Do we have some sort of equation similar to lamb-beer law for ELSD?

As AA and CPG have suggested, the answer is "maybe".

An ELSD essentially detects "dust" left behind when the solvent is evaporated from the droplets that come out of the nebulizer. To a first approximation, "dust is dust" and the response to each dust particle does not depend on the chemical structure of the non-volatile material. So, there is much less variation in response factors with ELSD than with UV.

That said, the presence of analyte may affect the number and/or size distribution of dust particles formed (e.g., by changing the surface tension of the column effluent or by affecting the rate of evaporation of solvent), so it's unrealistic to expect response factors to be exactly the same even for an isocratic separation (with a gradient of course, droplet formation may change quite a bit as a function of the mobile phase composition).

The relatively narrow (actually, nonexistent!) linear range is, to my mind, much less of an issue; I have no problems with appropriate mathematical transforms -- we do it all the time with UV detectors, which actually measure transmittance and then compute absorbance from that.

Bottom line: for "semi-quantitative" estimate, ELSD area% is probably a reasonable choice (certainly better than UV), but don't expect too much! :wink:

Re: is ELSD similiar to FID in terms of quantification

Posted: Tue Feb 21, 2012 4:28 pm
by tom jupille
Do we have some sort of equation similar to lamb-beer law for ELSD?
Unfortunately, none that I am aware of. :cry:

Re: is ELSD similiar to FID in terms of quantification

Posted: Fri Feb 24, 2012 2:34 pm
by jiangds06
Thanks all the same.
Do we have some sort of equation similar to lamb-beer law for ELSD?
Unfortunately, none that I am aware of. :cry:

Re: is ELSD similiar to FID in terms of quantification

Posted: Wed May 16, 2012 9:15 pm
by Sallybeetle
If I understand your question correctly, for ELSD we have the equations
y = Am^b and Log y = b Log m + Log A

Where:
y = Detector response (peak area)
A = Constant that changes with mobile phase and analyzed compound
m = Injected solute mass
b = Constant that changes with mobile phase and analyzed compound
[Generally, 0.67 - 2 depending on scattering mechanism (refraction-refraction to Rayleigh)]

The above equations can also be found in several reputable sources.

Re: is ELSD similiar to FID in terms of quantification

Posted: Thu May 17, 2012 12:59 am
by tom jupille
Yes, but those equations don't actually tell us very much; they say that the response "sort of" increases with increasing mass:
it's log-log, but the intercept can be pretty much anything and the slope can vary by a factor of 3 (and, in a gradient, both are likely to be different from one point on the gradient to another).