top board on 5972

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
How do you change the top board on a 5972?
It has been quite a few years since I replaced the top board on a 5972.

It was quite simple.
After venting, remove the cover and you will see the top board.
Disconnect any cables, and you will see several screws that hold the assembly.

Remove the screws and lift the board out.
Replace the board and the screws
Connect the cables.
Pump the system down

You must than dip the RF Coils.
I am a bit fuzzy on this since it has been so long.

You will see two threaded posts on the left.
You want to turn them as you monitor your progress in chem station.
I would contact Agilent and get advice on dipping the RF.
Overall the procedure was quite simple.

If you need to chat I would be available.

Regards
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry