Restrictions with 105m column on 6890?
Discussions about GC and other "gas phase" separation techniques.
11 posts
Page 1 of 1
We're currently running a 30m x .25mm x .25uM HP5-MS on our Agilent 6890 GC with a MS attached and considering using a 105m x .25mm x .25uM HP5-MS. Are there any foreseeable issues regarding flow, pressures, etc that we'll run into should we switch to the longer column?
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
You should do some calculations to check if you will have enough pressure in your capillary inlet. Agilent supplied two versions of the capillary inlet, one with a max pressure of 100 psi and the second had a max pressure of 150 psi. You should be able to check what you have installed by setting your 6890 to run in constant pressure mode, and then check what the max pressure that you enter is.
Gasman
Gasman
-
- Posts: 43
- Joined: Thu Jan 05, 2012 11:35 pm
It is the 100psi max inlet.
Where do I go from here?
Where do I go from here?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Use Agilent's downloadable free flow calculator software to see what column flow 90 psi head pressure (or whatever you choose) can deliver with your column dimensions.
-
- Posts: 43
- Joined: Thu Jan 05, 2012 11:35 pm
Inlet Pressure: 90psi
Inlet flow: 1.1mL/min
Split vent flow: 98.4 mL/min (20:1)
Outlet flow: 4.92mL/min
Average Velocity: 43.9cm/sec
Inlet flow: 1.1mL/min
Split vent flow: 98.4 mL/min (20:1)
Outlet flow: 4.92mL/min
Average Velocity: 43.9cm/sec
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Note Temperature. Gas viscocity increases with temperature. Thus higher pressuire is needed to drive a column in a warmer oven.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
time to consider hydrogen carrier.
Rod
Rod
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
chromatographer1
They may not have a choice, as our helium supplier has notified us that helium is going to be rationed in the US. From April until the end of the year we will only be getting 70% of our normal helium supply.
Gasman
They may not have a choice, as our helium supplier has notified us that helium is going to be rationed in the US. From April until the end of the year we will only be getting 70% of our normal helium supply.
Gasman
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
One question why do you need to go to a 105M when you use a 30M? You may do better to go smaller ID ( 0.18). You will use less He as column flow will be less and your resolution should improve.
-
- Posts: 43
- Joined: Thu Jan 05, 2012 11:35 pm
@Bigbear: Sort of an arbitrary order...we're using a 105m Rtx-5 on our GC and we want our GCMS column to match for conformity. I know retention times won't be exactly the same, but they'll be close.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Not as close as you might find out.
The difference of 15 psi over a 105 meter column might be a bit more than you anticipate. Of course you can always lower the flow rate on the MS column, or increase the temperature for the other GC.
I guess H2 carrier is out with the MS.
Rod
The difference of 15 psi over a 105 meter column might be a bit more than you anticipate. Of course you can always lower the flow rate on the MS column, or increase the temperature for the other GC.
I guess H2 carrier is out with the MS.
Rod
11 posts
Page 1 of 1
Who is online
In total there are 39 users online :: 0 registered, 0 hidden and 39 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 39 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 39 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science