Waters 717plus Autosampler
Posted: Fri Feb 10, 2012 10:08 am
Good morning all,
We have this autosampler in our lab.
One of my colleagues has been using it for quantitative work and found that he wasn't seeing any peaks at a certain injection volume.
I ran a check to see if the AS was consistent in terms of the volume taken from the sample vial. I filled a vial with water and after each injection (e.g. 100uL) I weighed to see if there was a loss of 100mg. Perhaps there is air trapped somewhere in the system, as the apparent mass taken up was much less than expected. It also varies depending on the position of the vial.
Is this a sensible way to assess the performance of the AS? This is quite an old design (pull-to-fill) and I'm no expert on them, but perhaps this check isn't suitable?
Any experts on here that can help?
many thanks,
Chrom
We have this autosampler in our lab.
One of my colleagues has been using it for quantitative work and found that he wasn't seeing any peaks at a certain injection volume.
I ran a check to see if the AS was consistent in terms of the volume taken from the sample vial. I filled a vial with water and after each injection (e.g. 100uL) I weighed to see if there was a loss of 100mg. Perhaps there is air trapped somewhere in the system, as the apparent mass taken up was much less than expected. It also varies depending on the position of the vial.