- Posts: 164
- Joined: Wed Jul 03, 2013 1:51 am
We've had terrible problems with phthalate contamination in our positive mode chromatograms. We see very abundant ions at 149 m/z (phthalic anhydride, fragment of 391), and 391,413,429 (diisooctyl phthalate +H,+Na,+K).
We replaced many parts on the LC and flushed and cleaned it very carefully but to no avail. We even moved the instrument to a different room to see if it was the lab air. We then used a different LC and the contaminant signal was almost exactly the same.
That's how we located the problem to our heated ESI source (HESI). Likely, the phthalate is evaporating from source parts other than the actual spray needle assembly, which was replaced with no ensuing change in contaminant level.
After starting the instrument, the signal from the ion at 391 m/z continues to increase over several 10s of minutes (long after the needle temperature readback has stabilized at 300C) until the whole source housing has equilibrated to a temperature that is hot to the touch. We will now clean the source housing and metal sheath encasing the spray needle.
This is a source of contamination is something I have never seen before and took a long time to pin down.
has anyone seen this problem? Where did it stem from? How did you clean the source and solve the problem?