Thermo Focus GC Problem

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I want to know about liners of Thermo Focus GC
Does it have 2 mm Splitless liner?
I am having trouble in analysis of API Residual Solvent as it says that blank (DMSO) should be analyte free and I am having methanol peak in it.
method parameters,
Det temp 200
Inj temp 140
gas Nitrogen

Please provide suggestion
The peak you are seeing is not methanol but it elutes where methanol elutes. (unless you have exposed your DMSO to methanol in the air or elsewhere) The peak may be dimethylsulfide.

You can pull vacuum on a portion of DMSO while warming and stirring it to remove volatile impurities

Nitrogen sparging would be helpful also.

If you heat a portion of DMSO try not to expose it to atmosphere as water vapor can contaminate your sample which can encourage reactions producing DMS.

Good luck.

Rod
will it be possible to use Rotavapour for it?
If yes then please tell me temp and other parameters like vaccum
3 posts Page 1 of 1

Who is online

In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry