Problem with a new column and its pressure
Posted: Tue May 03, 2005 12:40 pm
Dear members of this forum,
Yesterday I had a serious problem with the pressure in the column while I was injected. The pressure was 110 Bar and it started to increase until when it was 150 Bar I stopped; the mobile phase was Phosphate Buffer pH 7.2 with triethylamine (0.2 %) and Acetonitrile (50:50).
The column is Lichrospher RP-18 from Agilent. I changed the mobile phase to water and washed during 30 minutes (2 ml/min) at 60ºC (the pressure was below to 200 Bar) when I change to methyl alcohol the pressure increase to high values (upper to 200 bar) and I decided to stop. I do not understand because organic molecules were attached to this kind of column but the change of solvents behave against the polaritiy.
I would appreciate any help in order to restore the column because it is new, not more than two months. Other important point, I had worked with precolumn and I worked to pressure about 400 Bar.
Thanks in advance for your help,
Diego Delmonte
Yesterday I had a serious problem with the pressure in the column while I was injected. The pressure was 110 Bar and it started to increase until when it was 150 Bar I stopped; the mobile phase was Phosphate Buffer pH 7.2 with triethylamine (0.2 %) and Acetonitrile (50:50).
The column is Lichrospher RP-18 from Agilent. I changed the mobile phase to water and washed during 30 minutes (2 ml/min) at 60ºC (the pressure was below to 200 Bar) when I change to methyl alcohol the pressure increase to high values (upper to 200 bar) and I decided to stop. I do not understand because organic molecules were attached to this kind of column but the change of solvents behave against the polaritiy.
I would appreciate any help in order to restore the column because it is new, not more than two months. Other important point, I had worked with precolumn and I worked to pressure about 400 Bar.
Thanks in advance for your help,
Diego Delmonte