Flow rates on an Agilent 6890

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
Have a method for caffeine analysis and it says to use a flow rate of 10ml/min
this sounds high to me
Any suggestions guys
MD
Depends on the column , packed or capillary, id, etc.
It can also depend on the carrier gas. Using hydrogen allows higher flow rates.

Gasman
Hi Guys
Its a Capillary column (HP -17 from J&W Agilent) carrier gas is Nitrogen detector is FID
Thanks
What is the diameter of the column?

Depending on how easy the separation is you can go away from the ideal flow to speed things up, this could be one reason for the flow rate given. Or it could be wrong.
Where can I buy the kit they use in CSI?
Have a method for caffeine analysis and it says to use a flow rate of 10ml/min
this sounds high to me
Any suggestions guys
MD
Maybe that's the total flow; 1 ml down the column and about a 10:1 split ???

Peter
Peter Apps
Could be all of the above. I used about 10 cc/min with a 30M X 0.53 column. Worked fine for years. Switched to a 0.45 column and reduced the flow down to about 7.

Depends on the column ID.
thanks guy for your help
MD
8 posts Page 1 of 1

Who is online

In total there are 50 users online :: 1 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 49 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry