I am having trouble while using Porapack R for Residual Solvent IPA
The column shows decreasing peak response after longer use and RT also get changed.
How to tackle this Problem.
Ple help me out.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Porapak R and Q are both divinylbenzene porous polymers.
difference is surface area. The Q has much more retention and loadability, meaning you have to run it at higher temperature to get the components off. The Q is used most of the time.
If retention is changing it can be many things. Assuming that the flow is the same, and you see lower response, the column may get contaminated, blocking the pores. May use the for higher capacity, inject less or use a bit more sample cleanup. Also you may consider to use a liner with some glasswool, so you can replace that section periodically.