LC-UV detection of 5-fluorouracil (5FU)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello,
I am trying to develop a method on my agilent 1200 LC system for the analysis of 5FU with detection by UV. 5FU is in 0.1 M NH4OAc and the LC is running on an isocratic elution of 93:7 water:MeCN with a weakly retaining C18 column. On the low end of my calibration curve I am getting this dip in the UV chromatogram which I believe is causing my curve to follow a polynomial fit rather than a linear fit. My analyte peak is on the side of the dip at 2.286 min. Here is a picture

Image

I have seen this before in other UV chromatograms. Does anyone know what it is caused by? I am working on changing the chromatography to fix it now, but I am curious.
Thanks!
If you are using a diode array detector, check to see if the reference wavelength is set. If so, turn it off and see if that helps.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 185 users online :: 4 registered, 0 hidden and 181 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot], John Guajardo and 181 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry