I am using the Acquity C18 column 1.7 µm, 2.1*50 mm from Waters 2 years succesfully on our LC-MS/MS system. Normally, after 800-1000 injections (depending on the cleanliness of the samples, which is not evident since I work in a forensic lab), I have to replace the column because of bad peak shape, increased pressure etc.
I recently replaced an old column, and did 10 injection succesfully on the new column. Then, the pressure increased. The mobile phase I use is 10 mM ammonium bicarbonate at pH 9 and methanol, flow rate 0.5 ml/min, oven is at 40°C. The method is fully described in two of my papers: Verplaetse R and Tytgat J, For Sci Int (epub ahead of print) and J Chrom B 2010, 878, 1987-1996.
I flushed the column already with water, methanol, acetonitrile and isopropanol without improvement.
We have two similar LC-MS/MS- systems. On both systems the problem is seen and another new unused new column has normal pressure. This excludes a problem with the hardware or mobile phase and point directly to the colmun. I have no idea what the problem could be, but considering the price of these columns, I hope I can still solve the problem, since it was only used 10 times...
Any suggestions...?