Page 1 of 2
Conditioning sorbant in RP-SPE
Posted: Thu Jan 12, 2012 1:55 am
by DJ
Why must RP SPE material be conditioned with methanol, followed by water before use? If not conditioned, is binding or elution/recovery affected?
Re: Conditioning sorbant in RP-SPE
Posted: Thu Jan 12, 2012 2:58 am
by tom jupille
Yes, because there will be a good chance that the cartridge bed is not thoroughly "wet" (i.e., you won't have good contact between the bed and your sample).
Posted: Thu Jan 12, 2012 9:43 am
by HW Mueller
A long time ago I tried to improve HPLC of vasopressin via a SPE cleanup, using a C-18 cartridge. The vasopressin got hung up on the cartridge, so I gave up that cleaning step. A while later I bought vasopressin RIA kits that used exactly the same cartridge for the same purpose. It worked quite well, the only difference to what I did a few years earlier was a methanol washing step prior to sample application. I don´t know exactly why the peptide got hung up without the methanol washing step.
Posted: Thu Jan 12, 2012 1:24 pm
by Don Shelly
Was is the exact same C18 made by the same manufacturer?
Posted: Thu Jan 12, 2012 11:52 pm
by DJ
By "hung up" do you mean recovery was poor?
What if I'm not interested in recovery, only removing an impurity from an aqueous solution?
Posted: Fri Jan 13, 2012 7:03 am
by Peter Apps
I always thought that the methanol conditioning was to get the C18 phase "uncollapsed" - because the cartridges ship dry the phase is in the same state as a reverse phase column that has been run in 100% aqueous, the C18 chains are lying flat on the support and so do not have the capacity that they should.
Posted: Fri Jan 13, 2012 11:06 am
by HW Mueller
With "hung up" I meant that it stayed all on the cartridge and that I couldn´t get it off with a bunch of solutions (don´t remember what I used, one of them was about identical to what they used in the RIA kit).
I used the same cartridge as the kit manufacturer, although I don´t expect the batch having been the same.
Posted: Fri Jan 13, 2012 2:47 pm
by Don Shelly
The reason that I was asking is because you should not see variation in sorbent performance when using the same manufacturer. (The company that I work for manufactures SPE sorbents) If you do see variation, the manufacturer has a big quality issue.
You will see variation from manufacturer to manufacturer, because no two companies make it the same way or use the same raw materials.
I agree with Peter's explanation of what the methanol is for.
Don
Posted: Fri Jan 13, 2012 4:10 pm
by DJ
Interesting. Since polystyrene/DVB is a different animal than C18 altogether, I wonder if conditioning would have similar effect.
Posted: Fri Jan 13, 2012 4:14 pm
by Don Shelly
It helps, but it isn't as important as it is with silica based sorbents. The most important thing that it does is it ensures a clean product.
Posted: Sat Jan 14, 2012 10:18 am
by HW Mueller
Don, I got different results, because I didn´t wash with MeOH. When I did this the washing step was not yet "invented".
Posted: Mon Jan 16, 2012 8:09 pm
by DJ
Supposedly there are active sites onto which peptides irreversibly bind. It would be pretty discouraging, though, if this contributed to a measurable loss in recovery.
Posted: Mon Jan 16, 2012 10:57 pm
by tom jupille
Actually, it can, which is why you want to look at recovery when you are developing/validating a method.
Posted: Tue Jan 17, 2012 10:09 am
by HW Mueller
Well, recovery was 0% before washing with MeOH. My guess was that the peptide caked out on the stat phase. One sometimes sees this in TLC also.
Posted: Tue Jan 17, 2012 10:56 pm
by DJ
What is the origin of these so-called "active sites"? Are these hydrolyzed silanols?