baseline problem of HPLC
Posted: Fri Jan 06, 2012 9:07 pm
I am running HPLC samples use the same method and same sequence except I prepare a new mobile phase A which is 0.1% TFA in water. There appeared 3 huge peak in my blanks .
Another analyst reprepare the mobile phase, we changed mobliphase B and C ( ACN and IPA) and cleaned the system and aloso change another used column. but there is still something on my baseline. I realzied it might cause by water, so I used bottled HPLC water. the base line are not that bad, the noise peak decreased from 0.02 to 0.005 and they are not 3 peaks anymore but they changed to one with a small negtive peak there. Right now I am using good water, ACN and IPA try to clean my system, what else I can do?
how can I post picture?
By the way, the method was running all days long for few months and nevery has such problem. it is gradient from 90% TFA to 70% ACN with 5%IPA in 3.5 minutes and stay there for 1.5 minutes and back to 90% TFA . the column is a 10cm column and the detect at 278nm.
those peaks shows up at 2.5-3 minutes.
Another analyst reprepare the mobile phase, we changed mobliphase B and C ( ACN and IPA) and cleaned the system and aloso change another used column. but there is still something on my baseline. I realzied it might cause by water, so I used bottled HPLC water. the base line are not that bad, the noise peak decreased from 0.02 to 0.005 and they are not 3 peaks anymore but they changed to one with a small negtive peak there. Right now I am using good water, ACN and IPA try to clean my system, what else I can do?
how can I post picture?
By the way, the method was running all days long for few months and nevery has such problem. it is gradient from 90% TFA to 70% ACN with 5%IPA in 3.5 minutes and stay there for 1.5 minutes and back to 90% TFA . the column is a 10cm column and the detect at 278nm.
those peaks shows up at 2.5-3 minutes.