SAX for bromide

Posted: Fri Jan 06, 2012 5:14 pm
by push
Hi
I need to analyze bromide in water containing TriChloroBenzene residue. I use Parisil SAX (Whatman) column and KH2PO4 buffer as a mobile phase. Bromide is stronly interfered with the TCB residue. I am thinking to add an organic modifier to the mobile phase. Which column and mobile phase can you recommend?

Re: SAX for bromide

Posted: Fri Jan 06, 2012 6:24 pm
by tom jupille
That's a silica-based column, so solvent limitations are the same as those for any bonded-phase column. Adding either MeOH or ACN should make the TCB elute earlier. The other possibility is to change your buffer ionic strength. Increasing the salt concentration should make Br- elute earlier; decreasing it should make Br- elute later.