running 10 mMol sulfuric acid into a triple quad

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I want to run a biorad H form ion exclusion column into an lcms. I run with
10 mMolar sulfuric acid. Will that kill my source?

It's a Sciex 3000.
Yes.

Almost certainly. I once, in my tender youth, tried an even lower concentration into an Agilent source, on the grounds that sulphuric acid is volatile. What I forgot is that it's a strongly oxidising acid and doesn't need any help to oxidise even high-quality stainless steel components in the heat of a spray chamber.

The bits of the chamber that didn't go black, went green and gooey. Fortunately I noticed after only 30min, and after an embarrassed telephone conversation with Agilent (who spoke very slowly, a little bit LOUDLY, and in simple, easy words...) and a great deal of washing, the mass spec survived with only slight pitting of the spray cone. But I really, really wouldn't recommend it.

Gave beautiful sulphate adducts of the expected product, though!
I'd be careful about your HPLC components too.

I seem to recall reading about some sort of cartridge you put between the column and MS that exchanges out the strong acid.

Another option is to try and adapt the method to work with, say, TFA or HFBA. These are bad for ion sources too but not nearly as bad as destroying it.
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry