different retention time though the methods are the same

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Dear users,

I am seeking an answer of a strange phenomenon in my GC-MS results. I conducted an objective analysis using GC-MS towards a certain papaya variety extract to see the volatile patterns. The result showed that naphthalene occured at 30.45 minute. I conducted another test using the same method and column but different papaya variety. The result was naphthalene occured at different retention time: 37.12. What had happened?
One or both of the peaks that the MS search says is naphthalene, is not naphthalene Or a system fault, or operator error.

Peter
Peter Apps
Thank you, Mr. Peter. =]
Look for the match with the library ; is it higher for one of the compounds ?

Which software are you using ? ChromaTOF is known to find a lots of "false positives".

You really need to compare with standards to confirm compounds retention time.

Hope that helps,

Fée
4 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry