First of all I would not try to analyse for any atmospheric gas by using a syringe. Pipe it straight from the cylinder (via a regulator) into the GSV.
What flow etc. are you using for your column?
Have you run any standards? Do you know the retention times of some common peaks e.g. O2/N2, CO2, methane, ethane? According to Agilent Tech Note 228-387, water should elute between ethane and propane, closer to propane. It doesn't say what concentration of water they have in the sample though.
How often do you have to do this work? It might be easier to use something like a Shaw meter? Although for those you need a large amount of sample.
Hello,
We are preparing methods for determination several different gas mixes and to confirm +/-2% their contents. Also we need to determine purity of each gas used for preparing mixes accroding H2O presents. due to our customer ship gase mixes for lasers system they have control the level of h20 as impurity in each gas.
We use HP-Plot Q but it separate rather poor N2 and H2 and the separation of Ar and O2 is bad as well. We will start with Molsieve capillary column next week. We are using TCD but I realize it must be PD-HID for 1-2ppm of waters as LOD. Customer informed us with wrong LOD at 200ppm so we have chosen TCD.
We have some standard gas mixes ordered from AirProducts. The water standard is 500ppm in He.
We are purging directly GSV from cylider of our standard mixes by regulator connected with GSV sample inlet by teflon 1/8" tube (similiar like in hplc systems). The flow of filling sample is around 40-60ml/min. We have measured it. The CO2 determination in that way gave us very good linearity and precision. r2=0,999761 with crossing 0,0 point.
best regards,
JCzajka