GCMS PBDE 209 Fragment Ions

Posted: Wed Jan 04, 2012 8:50 am
by malang
n

Re: GCMS PBDE 209 Fragment Ions

Posted: Wed Jan 04, 2012 12:00 pm
by malang
n

Re: GCMS PBDE 209 Fragment Ions

Posted: Wed Jan 04, 2012 2:12 pm
by krickos
Hi

As I recall there are some published papers on that BDE-209 degrades even on GC column at 300°C over time, signifactly faster and earlier if a less inert column is used.
Hence shorter columns (like 15m) with thinner films of good quality is generally recommended to minimise residence time in column and elute BDE-209 before 300°C.

Pyrolysis seems like an odd technique for a flame retardant that is supposed to break down at elevated temperatures, but maybe it is a proven technique nowdays.