I am currently fighting (that's pretty much it...) with ChromaTOF (GC-MS-SPME method for aromas) to build my calibration (over 40 compounds) and I can get it to work correctly for all compounds.
Actually, the software can't find all the analytes and it seems to have a problem with the fact that peak width is changing troughout the chromatogram (I have very large peaks and trace components). So, when I am using a wide peak width (such as 30 sec - which is the case for some peaks), ChromaTOF can't find smaller peak OR it finds them but reject them as being out of the selected peak width. On the contrary, if I use a small peak width, such as 1-2 s (recommended by our sale representant), ChromaTOF can find 10 peaks within the same actual peak.
Iam getting to the point that I'll use two calibration, one for the large peak and one for the smaller peak... Is there another solution ? I already used the function "Peak broaden troughout chromatogram" (in the Data Processing Method form) but it didn't give satisfying results.
Another question I have is : where to set up the retention time window ?
Thank you so much for your help - I am "trying to like" ChromaTOF, but, as a former Agilent/MS user, I have hard time...
Fée