Hornet » Sun Jan 01, 2012 3:32 pm
Hi,
We have varain 450 GC coupled with 240MS Which has ion trap mass analyzer, Internal ionization.
I would like to clarify regarding 240MS Air/Water check.
The water level in the system is at "Elevated level". Mass 19/18 ratio is around 52% keeps fluctuating between 50 -60%
We did some routine and preventive maintenance to get rid of this.
We changed roughing pump oil,septa, Injector liner, Made a leak test with 1:1 Isopropanol and water. Baked out system after maintenance for 18 hours.
But the water level hasn't come down to acceptable level, Some time its not stable it shows its in acceptable level, Again the status message is change to elevated level.
Any suggestions are apprecieated
Thanks and regards,
R.Sundaresan
First, baking out for so long with air/water in the system is the best way to destroy your column, never do it again.
I can try to help you but i need some informations:
- before the bakeout you had both air and water at high concentrations in the system and after the bake out you have only high water level, this is what i understand from your posts, is it correct? If it's like this it sounds a bit weird to me, usually after bakeout the water level goes down to a mid/acceptable level even if there is a small to moderate leak, while the air level stays high. Please explain better this part.
- i need to know what is your injector model, is it a Varian 1177 split/splitless or a varian 1079? If it is a 1177, which o-rings are you using?
- where have you done the 2-propanol leak test? Just at column ends or somewhere else?