We are doing LC/MS method development using ABISCIEC 5500 MS with the Agilent 1200 series LC unit.
We are having issues with carryover and we have narrowed it down to the Agilent auto sampler. We believe the problem is leftover/residue in or around the needle. We will run two run blanks (40 ul) that show no peaks, then run our sample with the peak we are looking for, then another set of run blanks that show carryover with the peak from our sample. This peak gradually lowers in intensity with each run blank.
We have tried running the automatic needle wash/port using 20/80 acetic acid/methanol as the wash in the port. If I'm correct, this is supposed to clean the outside of the needle only. I have also tried running an extra line in the injector program to draw/eject this same wash before running the sample with no luck. This was done in an attempt to clean and rinse out the inside of the needle. I have also tried different combination of washes using methanol/water/acids/run blank solution.
The method we are running is a 20 minute gradient elution with A = water with 0.1% formic acid, B= acetonitrile with 0.1% formic acid.
I was hoping that this forum could provide some suggestions or any troubleshooting tips and hints. I would really appreciate the help.
Thank you.
We are having issues with carryover and we have narrowed it down to the Agilent auto sampler. We believe the problem is leftover/residue in or around the needle. We will run two run blanks (40 ul) that show no peaks, then run our sample with the peak we are looking for, then another set of run blanks that show carryover with the peak from our sample. This peak gradually lowers in intensity with each run blank.
We have tried running the automatic needle wash/port using 20/80 acetic acid/methanol as the wash in the port. If I'm correct, this is supposed to clean the outside of the needle only. I have also tried running an extra line in the injector program to draw/eject this same wash before running the sample with no luck. This was done in an attempt to clean and rinse out the inside of the needle. I have also tried different combination of washes using methanol/water/acids/run blank solution.
The method we are running is a 20 minute gradient elution with A = water with 0.1% formic acid, B= acetonitrile with 0.1% formic acid.
I was hoping that this forum could provide some suggestions or any troubleshooting tips and hints. I would really appreciate the help.