MS2 Scan Time

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello all,

I am attempting an MS2 Full Scan while trying to develop a method.

I would like to know about how to set MS2 Scan Time?

Right now the current setting is at 500?, Any idea as to what this number is or how to optimize?

I would like to scan from 700-800 AMU

I am using a Agilent 6410 QQQ, LC/MS/MS.

Thanks
it depends on what is your peak width and how many events you have in your method
if you have only one event, that is fragment with only one collision energy, then 500ms is tolerable, however I would suggest to decrease this to get more data accross the peak

when you set up several events in the method with different collision energies you should decrease the scan time so as to the total scan time would be adequate to the peak width, e.g. 5 events with 100ms scan time each
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry