Lower Peak Areas than Usual with More Than One Column
Posted: Thu Dec 22, 2011 1:09 am
I measure free and total glycerin in finished biodiesel using method ASTM D6584. I have an Agilent 6890N GC with an Agilent autosampler that injects a 1 uL sample (cool-on injection) onto a DB-5 HT capillary column. The detection is done by FID. I use 1,2,4-butanetriol as an internal std for free glycerin and tricaprin as an internal std for mono-, di-, and triglycerides.
About two weeks ago I began noticing that our QC samples were having trouble passing because everything was coming out at the high end of the std deviations (I use ASTM round robin biodiesel samples for QCs). Thinking that it was the fact that the column was about at the end of its lifetime, I decided to change the column out and replace it with an identical model from the same company. This did not fix the problem and I soon realized that the jet has a plug in it. I replaced the jet with a new one and thought everything would be fine.
I then calibrated the new column like usual with purchased std mixes and purchased internal stds. The calibration seemed to have worked fine with R-Squared values above 0.99 on the curves for each component. However, I began noticing that the peak areas of all components (especially the internal stds) were considerably lower than usual (about 880 for tricaprin compared to the usual area of about 1300 or so). This is in turn increasing our test results. I had a sample sent to two independent labs and I ran it as well to confirm that something was wrong. The two ind labs got similar (and normal) results for everything, but the numbers I got for results for each component were about 20% higher than the other labs' results.
I don't know if maybe I replaced a part (like the jet) incorrectly or what. I reinstalled the original column today and the peak area of the tricaprin std was down to about 850 or so from it's original area before removing of about 1300. I thought it was just a bad column after I put the new one in and saw the lower areas but now it is leading me to believe it is a problem with the GC at the inlet or the detector side.
Any help would be greatly appreciated!
About two weeks ago I began noticing that our QC samples were having trouble passing because everything was coming out at the high end of the std deviations (I use ASTM round robin biodiesel samples for QCs). Thinking that it was the fact that the column was about at the end of its lifetime, I decided to change the column out and replace it with an identical model from the same company. This did not fix the problem and I soon realized that the jet has a plug in it. I replaced the jet with a new one and thought everything would be fine.
I then calibrated the new column like usual with purchased std mixes and purchased internal stds. The calibration seemed to have worked fine with R-Squared values above 0.99 on the curves for each component. However, I began noticing that the peak areas of all components (especially the internal stds) were considerably lower than usual (about 880 for tricaprin compared to the usual area of about 1300 or so). This is in turn increasing our test results. I had a sample sent to two independent labs and I ran it as well to confirm that something was wrong. The two ind labs got similar (and normal) results for everything, but the numbers I got for results for each component were about 20% higher than the other labs' results.
I don't know if maybe I replaced a part (like the jet) incorrectly or what. I reinstalled the original column today and the peak area of the tricaprin std was down to about 850 or so from it's original area before removing of about 1300. I thought it was just a bad column after I put the new one in and saw the lower areas but now it is leading me to believe it is a problem with the GC at the inlet or the detector side.
Any help would be greatly appreciated!