I have a big problem with the determination of ethambutol hydrochloride and Meso HCl,
I take 0,2g of ethambutol in separating funnel and add 40 ml of NaOH solution and mix. Add 5ml of 0,1% 2,4 dinitro fluoro benzene in chloroform.
Extract with four quantities, each of 20 ml chloroform. Evaporating and dissolve the residue in 100 ml H2O/ACN 50:50.

HPLC system

column : Hypersil Silica 5µm
Mobile Phase : 20 ppm of CuSO4 5H20 in distilled Water :ACN 50:50
pH :7,0 adjust by very diluite Ammonia solution
Flow : 1,5 ml/min
Wavelenght : 254 nm

calculation :

Response factor of Ethambutol HCl and Meso HCl is almost same

ethambutol HCl : 1,00 Meso HCl : 0,97

Meso Hcl in Ethambutol HCl is calculated by area % method.

The problem is that in the chromatogram i get a single peak and have intensity very low to be a solution of 2000 ppm ethambutol.
I think that 2,4 dinitrofluorobenzene is a chiral derivating agents and extract only meso HCl......I don't know I'm not convinced by

Help me please
Thanks,
Antonio
First of all, DNFB is *not* a chiral derivatizing reagent. There is nothing chiral about it, it simply tags amino groups with a UV chromophore. That's why you get the same response factor for the two forms. Any enantioselectivity in this method comes from complexation with the Cu++ in the mobile phase.

The classical Sanger reagent method for amines indicates an absorbance maximum at 420 nm. 254 nm sounds suspiciously like a very, very, old LC method developed back in the days when fixed-wavelength 254 nm detectors were the norm. Unless you are running a validated method where you are forced to run it as-is, you might look at a different wavelength.

All of that said, I suspect that your derivatization reaction is not working properly. If this were my problem I would start with the ethambutal and establish that you are getting reproducible derivatization (look for a reasonably linear calibration plot). Once you have that, work on the separation.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hi!
Thank you very much for your answer,
I managed to improve the extraction with your invaluable advice...I'm very happy to have found this forum because is an important help for young analist!

Many thanks
Antonio
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry