Please let me know if there is a standardized procedure to follow to determine when do you need to change a column?

Is there a software that one can use to determine if there is a semnificative change in chromatogram that can compare an initial chromatogram with the current one to decide if the column needs to be changed.

Thank you

Re: When to change the column

Posted: Tue Dec 20, 2011 6:56 pm
by Consumer Products Guy
Change it when it can no longer do the function you need it to do. For cGMP, that might be resolution from another peak.

Re: When to change the column

Posted: Tue Dec 20, 2011 8:13 pm
by tom jupille
To expand a bit on CPG's post:

A validated method should have a set of "system suitability" criteria. Those might include resolution between key peaks, plate counts, tailing factors, repeatability, retention windows, baseline noise, . . . etc. For a regulatory method, the rule is simple: if you meet system suitability, your results are valid. If you don't, the results are invalid and cannot be used.

In most methods, as the column ages, the system suitability results will gradually change (retention times may drift, peaks will get wider, tailing will get worse, etc.). If you chart the results over time, you can extrapolate forward to get an idea when you will drop below acceptability and make plans to change the column before you get to that point.