How to determine Limit of Quantitaion (LOQ)
Posted: Mon Dec 19, 2011 5:39 pm
by isaac.lee
I am developing a GC method and I am trying to determine LOQ/LOD. What will be the proper steps to determine LOQ/LOD?
Re: How to determine Limit of Quantitaion (LOQ)
Posted: Mon Dec 19, 2011 7:51 pm
by DR
Step one - search these very forums for just your topic. Links to tutorials etc. have been posted before (a few times).
Posted: Thu Dec 22, 2011 12:53 pm
by wiclef
EPA proposed that the minimum detection limit is equal to the standard deviation limit multiply the student t value from statistical tables for 99% confidence level and n-1 degrees of freedom.
The minimum level of quantification ML is equal to 10 multiply to the standard deviation but normally we calculate the ML from the MDL.for example a sample size of 7, the MDL = 3.143 * S so ML = 10 MDL/3.143=
3.18 * MDL.
I think that I give you the answer so try to do this.In addition, the procedure is not harmonized so you judge yourself.
Wiclef Kagisha
Posted: Thu Dec 22, 2011 1:29 pm
by Hollow