Tributylamine
Posted: Mon Dec 19, 2011 3:49 pm
by dsanto01
does anyone have any advice/protocol for removing tributylamine (m/z 186) from a nanoLC/QToF system?
many thanks
Clive
Re: Tributylamine
Posted: Tue Dec 20, 2011 4:14 am
by mckrause
Are you sure it's tributylamine? You would have to have had it in significant quantities.
Use an acidic washout solution - it'll help dissolve the TBA and carry it out of the system.
Re: Tributylamine
Posted: Tue Dec 20, 2011 5:17 am
by tom jupille
Re: Tributylamine
Posted: Tue Dec 20, 2011 12:20 pm
by sav
many thanks
Clive
Trace of TBA (186amu in +ESI) is very difficult to completely remove (after working with TBA).
1. Wash your HPLC with 0.1%v FA in water, then 0.1%v TFA in MeCN, then MeOH and water (approx. 50-100ml each). In some cases, you need to replace spare parts (degaser, capillaries, in-line/inlet filters etc.).
2. Wash injector/AS. In some cases, you need to replace spare parts (seals, syringe etc.).
3. Wash ion source.
P.S.: I don't use TBA anymore :)
Re: Tributylamine
Posted: Tue Dec 20, 2011 1:54 pm
by dsanto01
Thanks for all the advice. We were using tributylamine as an ion-pairing reagent for the analysis of nucleotides, not knowing the sticky nature of the compound.
will try the advice and report back.
thanks
Clive