Limit of detection and area under the curve

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,

Please let me know if I can use the Mean + 3xSTD to determin the LOD in case of an HPLC where I evaluate the area under the curve. In this case how do I do to measure the area under the curve for the baseline noise?

Please help me with suggestions in this case.

Thank you very much to all of you
Step 1: search through these forums and you will find discussions on determination of LOD and LOQ.
If that does not do it, I would suggest looking at the US EPA methods - they are online and free. they have rules for finding limits of detection.
One advantage to the "statistical" approach (based on the standard error of the y-intercept) is that you don't have to measure the noise.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry