Karen01 » Sun Dec 18, 2011 5:20 pm
It is common to use the total ion chromatogram from a GC-MS (at least in EI mode) like the FID trace for estimating composition of a mixture. And, there are discussions of the validity of such measurements on this forum.
I did a search on "total ion chromatogram" in this forum and could not find the discussions... Could you provide links to either inside the forum or out ?
Update: I did find one:
viewtopic.php?t=9944&p=47482
After reading that I came away that doing it by FID carbon counting my error would likely be less than 10%. I'm still not sure how much potential error there is likely be trying to use an EI TIC like an FID in terms of # of carbons.
The assumption that the total ion chromatogram would give responses proportional to concentrations of analytes would imply that we are assuming the same ionization efficiency, perhaps increasing with the size of the molecule, for all compounds (at least within a class) and that we count all the ions.
The carbon number number can vary by 7 from my 'standard' ... We are talking about chains in the 20s.
Given how easily hydrocarbons fall apart, for TICs there would obviously be more ions for larger molecules, but I would expect branching could impact that number by affecting fragment sizes because of ion stabilization, though i don't know how significantly.
Has there been any work done to correlate # of carbons (thus MolWt) to TIC response for hydrocarbons? For unbranched I would think there could be a decent relationship. Perhaps that was discussed in the the threads you are referring to? It would be nice if the TIC peak area was proportional to Mol Wt (or carbon number) in the class.
If the ion production differences are not big in the range I'm looking at, or is directly proportional to carbon #, then perhaps I could just quantitate against the single standard I have on a mass basis. If not I would need a relationship between Carbon # and TIC area.
Of course to use TICs as with FID, I also have to separate the compounds from everything else... in that case, if I can do that, as I said the most straightforward way is to ID the peaks by MS using the parent ion and quantitating by FID.
If you want to look at the molecular ion only, you would have to assume that it is formed in the same proportion for all compounds. Examination of the spectra of the normal hydrocarbons, for example, shows a fairly healthy molecular ion for the smaller hydrocarbons and a very small intensity molecular ion of the larger hydrocarbons. And, you see more and a greater number of fragment ions being formed in the larger hydrocarbons. This suggests that the response factors for molecular ions will vary widely – and will not be useful without some form of calibration.
Does that not imply the same thing for TIC peak areas unless they are directly proportional to carbon #?
Thanks,
- Karen