GC or HPLC with glycols
Basic questions from students; resources for projects and reports.
9 posts
Page 1 of 1
When you have a mixture of glycols, glycerin and water, which do you think the best chromatographic method to identify the glycols?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Either could work. GC would probably require derivatization, which is a pain in the nether regions. LC would have detection limit issues (no chromophore).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
Pulsed amperometric detection works perfect for glycol and similar alcohols.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
GC with trimethylsilyl derivatization, non-polar DB-1 or DB-5 capillary. Dilute with DMF, transfer some to autosampler vial, mix in derivatizing agent, cap and shake, inject into GC. The reagent will also react with any residual water, so make sure to have excess reagent.
This has worked well for me for 30 years, helped put my two kids through college.
This has worked well for me for 30 years, helped put my two kids through college.
-
- Posts: 2
- Joined: Fri Mar 16, 2012 3:14 pm
Hi cpg,GC with trimethylsilyl derivatization, non-polar DB-1 or DB-5 capillary. Dilute with DMF, transfer some to autosampler vial, mix in derivatizing agent, cap and shake, inject into GC. The reagent will also react with any residual water, so make sure to have excess reagent.
This has worked well for me for 30 years, helped put my two kids through college.
about Deg (diethilene glycol) and trimethylsilyl derivatization what're the best gc-fid conditions? I m trying but can't see the peak of derivate product. My temperature set go from 80ºC to 300ºC (inj: 300ºC) , end time is 30 minutes with a 30meters non-polar column. Any help? Thanks!
-
- Posts: 102
- Joined: Mon Jan 10, 2011 1:38 pm
Glycols and diols can be easily done with GC. Derivatization is not required. Best selectivity and peak shape for diols is obtained using wax columns like Rtx-Wax. If water is the matrix, the injection will be the biggest challenge. Glasswool will be a challenge, so best use a liner with higher surface like cyclosplitter. Some like the inverted cup. Double gooseneck works, but reproducibility is not that good. If you can use hot-needle injection, use that, as it's the best for such samples.
To reduce the impact of (big)water peaks on chromatography, try to do the analysis at higher temperatures, using thicker Wax films. 1 micron on 0.32mm ID will do.
Jaap
Restek corporation
To reduce the impact of (big)water peaks on chromatography, try to do the analysis at higher temperatures, using thicker Wax films. 1 micron on 0.32mm ID will do.
Jaap
Restek corporation
-
- Posts: 2
- Joined: Fri Mar 16, 2012 3:14 pm
Thank you jdezeeuw, very interesting! Matrix is seawater so like you said implentation of a direct injection maybe hard and i need to use the gc we have "without spend too much time or money" (my boss words, so no new column). So i was trying derivatization, we have only 2-3 samples every year so i don't mind losing few hours in it. Only problem is dryness the sample before derivatization. There is something else that i can use in aqueous solution?
P.s. This morning i see my peak!
very small but it made my day!
P.s. This morning i see my peak!
-
- Posts: 95
- Joined: Mon Feb 12, 2007 10:05 pm
Hi Jaap,
Can you point me towards some more information on the Restek non-derivatized method? I'm looking for ethylene and propylene glycol in aqueous samples-but will be running several hundred per week.
Thanks,
B
[quote="jdezeeuw"]Glycols and diols can be easily done with GC. Derivatization is not required. Best selectivity and peak shape for diols is obtained using wax columns like Rtx-Wax. If water is the matrix, the injection will be the biggest challenge. Glasswool will be a challenge, so best use a liner with higher surface like cyclosplitter. Some like the inverted cup. Double gooseneck works, but reproducibility is not that good. If you can use hot-needle injection, use that, as it's the best for such samples.
To reduce the impact of (big)water peaks on chromatography, try to do the analysis at higher temperatures, using thicker Wax films. 1 micron on 0.32mm ID will do.
Jaap
Restek corporation[/quote]
Can you point me towards some more information on the Restek non-derivatized method? I'm looking for ethylene and propylene glycol in aqueous samples-but will be running several hundred per week.
Thanks,
B
[quote="jdezeeuw"]Glycols and diols can be easily done with GC. Derivatization is not required. Best selectivity and peak shape for diols is obtained using wax columns like Rtx-Wax. If water is the matrix, the injection will be the biggest challenge. Glasswool will be a challenge, so best use a liner with higher surface like cyclosplitter. Some like the inverted cup. Double gooseneck works, but reproducibility is not that good. If you can use hot-needle injection, use that, as it's the best for such samples.
To reduce the impact of (big)water peaks on chromatography, try to do the analysis at higher temperatures, using thicker Wax films. 1 micron on 0.32mm ID will do.
Jaap
Restek corporation[/quote]
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
With the information given I would prefer to use HPLC if I were doing the analysis.
But good luck in any case.
Rodney George
But good luck in any case.
Rodney George
9 posts
Page 1 of 1
Who is online
In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science