Hornet » Fri Dec 16, 2011 11:32 pm
Baseline is subtracted and it's thankfully not picking up a point on the peak. 442 tends to be base peak at any concentration level. After 10ng on column it is about the only ion you see in the spectrum.
Ion-trap temp of 150
Manifold temp of 120
Transferline temp of 150
Solvent MeCl2
Refer to EPA 625 for our list of analytes
Your peaks are tailing because your spectrometer temperatures are way too low for the analytes on that list. Remember that the transferline temperature must be at least as high as the highest oven temperature reached by the method (in your case 340°C), it's better if you set it at 20°C more if the column specs permit it.
Your trap temperature is too low as well, i would try to increase it at least to 200°C and see if it works better, you can raise it to 240°C without any problem for your instrument (the only problem would be low sensitivity for low-boiling compunds but if the instrument is dedicated this isn't a issue), and remember that at 250°C (highest temp. permitted) the silchrom on the electrodes will start to deteriorate.
The manifold doesn't require to be that high, default temp for analysis is 50°C but if you raise the trap at 200°C+ it will never reach 50°C, so you better put it at 70°C like i did and end of the story.
2 microl of methylene chloride on a 1177 splitless liner will overload your injector (from FlowCalc) if you don't put any additional pressure, resulting in loss of analytes/sensitivity, backflash, etc. 30 psi will be more than enough for a pulsed splitless pressure with that inj. volume.
So: try higher spectrometer and transferline temperatures, and try a pulsed splitless at 30+ psi.
IMPORTANT: i don't know how many injections have you made with these instrumental conditions, in my opinion the last 10-20 cm of your column end (xferline end) are heavily contaminated with analytes that didn't make it through the xferline due to low temperature. Before trying any of the above solution you must turn off the instrument and cut the column to prevent trap contamination.
If you still have resolution and tailing problem after all this operations your trap is probably contaminated because of the low temperature that overloaded it with heavy compunds, at that point you have to decide (basing your opinion on the number of injections you did) if a bakeout is enough or if you have to manually clean the electroes. Since you said you only injected standards and no field samples i think a bakeout should be enough.
Good luck.