Steroids/Hormones: transfer problem between instruments
Posted: Thu Dec 15, 2011 10:18 am
I prepare three standard solutions (80%, 100% and 120 % level) and three solutions of my sample. I fill each of the six solutions in two vials resulting in Vial series 1 and vial series 2
I make a sequence with vial series 1 on my HPLC instrument No. 1 using Column No. 1 and Mobile Phase Batch No. 1:
r2: 1.0000
y-intercept: 0.0
Assay of sample: 100 %
6 hours later, I take column No. 1 and mobile phase No. 1 and connect both on HPLC instrument No. 2. Then, I make a sequence with the vial series 2:
r2: 1.0000
y-intercept: 0.0
Assay of sample: 93 %
What is wrong? The different instruments or the different vial series?
=> Sequence with vial series 2 on HPLC instrument No. 1 and Mobile Phase Batch No. 1 and Column No. 1:
r2: 1.0000
y-intercept: 0.0
Thus, the difference in assay is based on the different instrumentation only.
Did anybody of you ever had such problems with steroids / hormones? How did you solve the problem?
Thank you very much for every hint.
Florian
r2: 1.0000
y-intercept: 0.0
r2: 1.0000
y-intercept: 0.0
r2: 1.0000
y-intercept: 0.0
Florian