Hello together,

could someone please tell me how to determine system dispersion the easiest way?

Thanks in advance!
Hi Anja,

have a look at a column "Care and Use" manual from Waters (e.g. XBridge Columns, Docu: 720001366EN). There it is such a procedure described
(look also for band broadening/band spreading etc.)

In general:
- replace column with union
(- add some pressure device after detector (or between pump and Injector))
- set the data rate as high as possible
- use a low flow rate (e.g. 0.5 ml or below)
- inject a small volume (1-5µL) of a test mix of enough concentration as to get enough absorbance (e.g. coffein, uracil, parabens or similar compounds)
- evaluate the peak width at 4.4%
2 posts Page 1 of 1

Who is online

In total there are 298 users online :: 0 registered, 0 hidden and 298 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 298 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry