If you use mole sieve available commercially (with the exception of Supelco's specially prepared MS 5A, if they still sell it, that is) you must use subambient oven temperatures to achieve separation of O2 and Ar.
Some thick film MS5A capillary may be capable of the separation, but when the NIST asked me to prepare them a packed column that would work at ambient temperatures I sent them a 9 meter 2.1mm ID column containing a special preparation of MS 5A.
I don't know if Supelco continued to make this packing or not.
I can send you a chromatogram example of air producing three peaks (O2, N2, and Ar) if you send me your email address.
best wishes,
Rod