Perkin Elmer GC-ECD, baseline and noise problem
Posted: Tue Dec 13, 2011 3:44 pm
Hello,
I have a Perkin Elmer GC-ECD instrument. I am trying to measure nitroglycerin (NG) in soil. I use the below method:
Injector temp: 150 oC
Oven Temp: 35 hold for 3min, 250 hold for 1 min
ECD temp: 300
SETPOINT 1.50
Carrier gas: He
Column: 30m, elite-5 , 0.25, 0.25
I used to get good results with the above settings. Starting 2 weeks ago I noticed that the base line goes up when the temperature reaches about 110-140. NG peaks comes out in 15.2 min (temp:~ 155). I also get false positive and noise. When I use only alcohol sample and injcet it I get noise and false positive. I already replaced the septum, replaced the injector liner, washed and cleaned the syringe, Backed the column at 320 for 2hrs, cut about 6 inches of both end of the column and baked it again. Nothing worked. Now even I don't get the noise. If I do not inject anything or inject alcohol I get flat base line until the temperature reaches 110 and then the base line goes up to 1000mv and saturate and keep going straight at 1000mV.
PLEASE HELP ME
Thanks
I have a Perkin Elmer GC-ECD instrument. I am trying to measure nitroglycerin (NG) in soil. I use the below method:
Injector temp: 150 oC
Oven Temp: 35 hold for 3min, 250 hold for 1 min
ECD temp: 300
SETPOINT 1.50
Carrier gas: He
Column: 30m, elite-5 , 0.25, 0.25
I used to get good results with the above settings. Starting 2 weeks ago I noticed that the base line goes up when the temperature reaches about 110-140. NG peaks comes out in 15.2 min (temp:~ 155). I also get false positive and noise. When I use only alcohol sample and injcet it I get noise and false positive. I already replaced the septum, replaced the injector liner, washed and cleaned the syringe, Backed the column at 320 for 2hrs, cut about 6 inches of both end of the column and baked it again. Nothing worked. Now even I don't get the noise. If I do not inject anything or inject alcohol I get flat base line until the temperature reaches 110 and then the base line goes up to 1000mv and saturate and keep going straight at 1000mV.
PLEASE HELP ME
Thanks