WHAT IS NEED TO CALCULATE RRF IN HPLC ANALYSIS
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
No need to shout
General way is to determine calibration lines of the reference substance (in pharmaceutical analysis usually the API) and the analyte(s) (usually related substances). RRF is calculated by dividing the slopes of the calibration lines (slope reference/slope analye if I remember correctly, but it might be the other way round...).
Another way would be to determine the responses (aka peak areas in HPLC) and calculate the RRF directly from it. But you should do this at several concentrations covering the working range to make sure the RRF is constant. Since you should use several concentrations anyway, the "calibration line approach" depicted above would be my way to go...

General way is to determine calibration lines of the reference substance (in pharmaceutical analysis usually the API) and the analyte(s) (usually related substances). RRF is calculated by dividing the slopes of the calibration lines (slope reference/slope analye if I remember correctly, but it might be the other way round...).
Another way would be to determine the responses (aka peak areas in HPLC) and calculate the RRF directly from it. But you should do this at several concentrations covering the working range to make sure the RRF is constant. Since you should use several concentrations anyway, the "calibration line approach" depicted above would be my way to go...
-
- Posts: 2
- Joined: Wed Nov 16, 2011 2:21 pm
BUT WHY RRF IS NEEDED ?WHAT IS NEED TO CALCULATE RRF IN HPLC ANALYSIS
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Different compounds have different response factors.BUT WHY RRF IS NEEDED ?

-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 71
- Joined: Fri Jul 08, 2011 4:48 pm
At times in the case of an impurity or related substance, you may not be making a standard and calibration curve of the compound. Instead you reference it's area in relation to the API, and then using the RRF can calculate out the concentration of the peak.
5 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science