I need to use UV detector to record arborption at 234 nm. One of my solvents contains some chloroform. Supposedly chloroform absorbs heavily at this wavelength. Any suggestions what solvent to try instead?

What kind of solvent system (with non-chorinated solevnets) would you recommend for resolving lipid classes by normal-phase HPLC?
One of the problems with normal-phase in general is that many (if not most) suitable solvents have UV-cutoff issues. My first suggestion would have been methylene chloride. Other than that, you can pick something of similar polarity from the list:
http://murov.info/orgsolvents.htm#TABLE%202
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry