625 & Varian 3800
Posted: Thu Dec 08, 2011 10:12 pm
Problem: Pentachlorophenol and Benzidine response at 5 to 10% of DFTPP/DDT and tailing is above method spec. Loss of spectral integrity beyond 50ng on column for DFTPP and non-active compounds.
System
Varian 3800 GC w/ 2200 Ion-trap MS
1177 split/splitless injection port
RXI 5SIL-MS with 10m integra guard 30m x 0.25 x 0.25
4mm ID single-gooseneck Restek Sky Liners; necked end @ column end
Run program:
Oven start: 40C
Oven ramp: 40C 1minute to 340C @ 10C/min
Injector: 250C
Flow: 1ml/min Ultra UHP Helium w/ clean background spectrum
Split: On @ 0.15min 50:1
Injection volume: 2ul
Column, liners, nuts/ferrules and injection port are new. Column is low bleed. No field samples have been processed. So I'm thinking method optimization. Anyone have experience with these Varian GCs and running SVOC by ion-trap?
Thanks!