unmgvar » Thu Dec 08, 2011 7:07 pm
probably you can find a newer column instead of lichrosphere that will work very much the same for most applications
i would go for something of today with 15 or 17% carbon load, since today columns are end capped and with low metal content
for the select B things are a little more complicated. this column had very poor batch to batch reproducability.
today ,merck to solve this sale their master batch select B.
it was their first attempt at creating a low metal content column, by performing some sort of wash to the silica. this was before the type 2 columns came out.
it is a C-8, with a very small pore volume of 60A, but a very small surface area on comparison, with 11.5% carbon loading
at first i would try a standard C-8 with about the same surface area and about the same C% loading.